I have to validate a method for residual solvents analysis, with an analytical range of 60 (from 0.1% to 6% of solvent content).
I can choose between a "single-point" calibration (three standards at a single concentration level) or a three level calibration (three standards ranging within the quantitation range).
Which one should I choose? I don't feel "comfortable" with the single-point, but, actually, both methods are linear in the analytical range.
Can anybody tell me which is the most commonly accepted practice with this kind of analysis?
![]()
![]()
![]()
![]()
By Anonymous on Thursday, September 12, 2002 - 11:20 am:
It is my understanding that if you have linearity and near zero intercept on the x-axis, a three point calibration is preferable and less likely to be contested by a reviewer.
I would use a point at or below the lowest level of measurement and one above the highest level of measurement. But a question, why only 3 points, why not 5 points. Two outside the measurement range and 3 points within sounds even better to me.
![]()
![]()
![]()
![]()
By Ron on Thursday, September 12, 2002 - 02:36 pm:
If you are doing a USP type of residual solvent analysis the most common method, 467, calls for a single point calibration. Without getting into the overall quality of the method, it is set up as a basic pass/fail situation. If all you are doing is pass/fail, then a single point calibation at the action level is all you need, assuming reasonable instrument stability. If you want to do more than a simple pass/fail test, then a multipoint calibration is better. 3 vs. 5 vs. more is an interesting statistical question, but for a less than 2 order of magnitude range 3 points will do fine, again assuming linear response over the calibration range.
![]()
![]()
![]()
![]()
By Anonymous on Thursday, September 12, 2002 - 09:09 pm:
1000 ppm to 6% is certainly not a USP test.
I believe the poster desires an accurate measurement of the solvents which may be necessary to determine std purity of a drug or a related, or to assist in process improvement.
If you want confidence in your determination and its application to whatever use you desire demonstrate not guess its actual content. That requires a high recovery below the lowest level of measurement and a demonstration of linearity through and past the high measurement point which should have 3 or more additional points. A 5 point regression as I posted above is performing a thorough and nearly uncontestable determination. Do the points curve? Do you recover 90% of an additional 1000ppm spike at 1000ppm level of contamination? Do it right the first time so no one can contend with you about your results. IMHO
:-) But you can always do it over. Job security and all that.
![]()
![]()
![]()
![]()
By labcat on Wednesday, September 18, 2002 - 02:06 am:
Thank you all for your answers.
I actually need to perform a quantitative analysis, not a simple limit test, that is the reason for my preference for a 3 point calibration. I wanted to be sure that somebody other share my opinion: now I feel more "comfortable"!
I agree that 5 standards would give narrower confidence limits on the results, but I have to make a compromise between the number of samples and that of standards.
Posting is currently disabled in this topic. Contact your discussion moderator for more information.