I need help with the USP Glycerin test for the limit of Diethylene Glycol. I'm running on a Agilent 6890, split with 10:1 ratio. Water based solutions. 0.05mg/ml std concentration. OVI-G43 capillary column, 30m - 0.53mm, 3micron film. Linear velocity flow of 38cm. Problem is that I have a terrible ramp covered with multiple peaks coming off for the last two thirds of the run and my peaks of interest are not showing up. They show at higher concentrations(.5mg/ml) Any ideas about this would be appriciated. Thanks.
![]()
![]()
![]()
![]()
By Soap Guy on Thursday, November 13, 2003 - 11:50 am:
Mike, I work for a major U.S. soap company, and we make our own USP grade glycerine, so we (R&D)had to incorporate this new USP test for our production facility. Like some other USP test methods, this one is not well-written, so as usual we re-wrote into our own format. The only thing I could figure out why and when the USP reference glycerine would be needed would be to establish a retention time, and maybe for the resolution solution. Anyway, we worked this out at R&D, and transferred to production QC where it is now used. I'm going try to upload some sample chromatograms in a later posting. We did find it important that the inlet liner be of an inverted cup type (Restek #20990 mini-Lam split liner) or a spiral type (Restek #20706 cyclosplitter). We are able to see peaks of the 0.05 mg/ml DEG.
![]()
![]()
![]()
![]()
By Soap Guy on Friday, November 14, 2003 - 07:35 am:
OK, here's some chromatograms. I don't know if the annotations will show up so first is dilute resolution solution of DEG and glycerine, second is 0.05 mg/ml DEG, and third is actual glycerine sample. The X-axis may be slightly different on these, go by retention times. Post back next week if you're still having problems.
res2.cdf (26 k) |
deg.cdf (26 k) |
glyc.cdf (26 k) |