Hi,
Working on CO2/CO/CH4/H2 system with Ar carrier gas. Did calibration manually (i.e. recording peak areas at different vol% of gases). Unfortunately, they don't add up to 100% when doing experiments (sometimes less than 90%, other times exceeds 120%). Any suggestion why? Thnx in advance.
![]()
![]()
![]()
![]()
By Anonymous on Monday, May 3, 2004 - 05:52 am:
It appears that linearity over the range of measurement is not achieved. Narrow the range of measurement for each calibration mix and your errors of measurement should then be minimized.
![]()
![]()
![]()
![]()
By Les on Saturday, May 8, 2004 - 05:24 pm:
How exactly are you injecting your gases. Putting the sample under pressure, you'll get more gas in the GC. I inject gases through a needle valve and bubble the vent into water. When the bubbles slow after sufficient purge, I turn the sample valve. If you sample while the bubbles are flowing, the charge is high due to more gas in the loop. Generally I shoot for a charge of about 85 to 110 and normalize the concentrations. If samples charge above the recommended, there may be problems due to areas exceeding the calibration range.
![]()
![]()
![]()
![]()
By Anonymous on Wednesday, May 26, 2004 - 08:44 pm:
check all the valves and the connections, it might be leaked. rethighten if posibble.
a big shift in total area/conc reproducibility almost always be due to:
1. a leak in the carrier gas system(valves and lines)
2. a problem with the detectors
3. a change in the injection technique
check all those items, its may helps.
rgds
omar.