Hi,
Iam screening carbamate pesticides on an API2000 with an 1100 Agilent LC. My column is a phenomenex Luna 3u C18(2) 150 x 2.0mm. My mobile phase is A=Water w/ 10mm Ammonium Acetate and .1% Formic Acid. B=Acetonitrile with 10mm Ammonium Acetate and .1% Formic Acid. My current gradient is A/B (95:5) to A/B (5:95) over 20 min, to A/B (95:5) in 2 min, then a hold at A/B (95:5) for 5 min. I am injecting a 10 ul sample injection which is in 80:20 MeOH/Water. I am unable to reproduce retention times and area counts. Any thoughts or suggestions? Also I am finding that with this gradient my pesticides are not eluting until 10 mins and later. Whereas phenomenex's elutes at 5 mins under the same conditions just a little bit different gradient. Instead theirs is A/B (90:10) to A/B (30:70) in 20 mins to A/B (90:10) in 2 mins then hold 5 mins at A/B (90:10). Any thoughts or suggestions on how to reproduce the same chromatogram as phenomenex's or close to it without turning some peaks into short, broad, jagged peaks? Thanks
![]()
![]()
![]()
![]()
By MG on Thursday, December 18, 2003 - 06:47 am:
On the retention time issue, your re-equilibration time at the end seems a little short for that length of column, assuming you are running in the 0.2 - 0.3 mL/min flow rate range. Try 10 min.
If there are problems specifically with your early eluting peaks, especially if they look ugly (fronting, split), try a weaker injection solvent. If the injection solvent can't be changed, try a smaller injection volume.
Is Phenomenex using the same buffers as you? Same flow rate? If so, try starting with their gradient and change from there to get what you need.