I am unable to reproduce my area count from one standard to the next, with the same standard concentration. I have changed metering seals and rotor seals on the LC. My retention times are consistent, yet the areas are not. Any suggestions? Thanks
![]()
![]()
![]()
![]()
By Uwe Neue on Wednesday, January 7, 2004 - 06:13 pm:
Sounds like injection problems, including air bubbles, tight seals etc. Check previous discussions of similar issues!
![]()
![]()
![]()
![]()
By Anonymous on Wednesday, January 7, 2004 - 07:02 pm:
How bad are they, MS is not quite as reproducible as you might expect from UV.
![]()
![]()
![]()
![]()
By MG on Thursday, January 8, 2004 - 07:30 am:
I've never run an API-2000 as the original poster has, but if it's any help, our API-3000 usually had %RSD's ranging from 2% to 8% for reinjection of a solvent standard, depending on the method and signal-to-noise at the level being injected. When it started creeping up to 8% and above, it was time to clean the source. I recommend cleaning the atmospheric pressure side of your source if this hasn't been done in awhile. Are you using ESI or APCI?
If source cleaning doesn't help, I would also suspect a problem with the autosampler.
![]()
![]()
![]()
![]()
By Anonymous on Thursday, January 8, 2004 - 09:44 am:
As a quick+dirty check of reproducibility, make replicate injections and weigh the vials before + after, if you are injecting enough to weigh accurately. If your weight RSD is much lower than the area RSD, look at detector. If weight RSD matches the area RSD, look autosampler.