Large baselinedrift with refractive index detector

Chromatography Forum: LC Archives: Large baselinedrift with refractive index detector
Top of pagePrevious messageNext messageBottom of pageLink to this message  By Christophe Courtin on Thursday, October 26, 2000 - 03:06 am:

We have been experiencing very large linear basline drift with our VSD Optilab refractive index detector. System is 0.3% NaCl MQ water at 0.5 ml/min over a HPSEC Shodex SB-804 column with Kontron pump. Drift is order of 100 mV in 30 min. The supplier's technicians don't find the problem. (UV is working fine)

Can somebody comment on that ?


Top of pagePrevious messageNext messageBottom of pageLink to this message  By hplcman on Friday, October 27, 2000 - 05:21 am:

Most drift with RI detectors is caused either by temperature or a slight leak in the reference valve. I would certainly check out the temp issue first.

Have you looked into a temperature issue yet? Does your lab experience shifts in temperature from morning to afternoon? Does your detector have an internal temperature controller?


Top of pagePrevious messageNext messageBottom of pageLink to this message  By Christophe Courtin on Friday, October 27, 2000 - 08:25 am:

The RI detector doesn't have a temperature controler.Temperature is an issue at some times (especially during warmer summers), but drifts are never near as large as they are now. We'll check the reference cell again, although it looked ok at first sight.


Top of pagePrevious messageNext messageBottom of pageLink to this message  By hinsbarlab on Thursday, November 2, 2000 - 05:13 am:

In the absence of a leak in the reference cell, another common cause of baseline drift is that the RI of the mobile phase is changing relative to that of the solvent in the reference cell. We always degas our mobile phases by helium sparging then maintain a helium pressure of about 5 psi in the mobile phase container to prevent reabsorption of air. You will need to purchase an apparatus designed for this purpose with an appropriate safety valve. I believe Alltech sells one. Inadvertent overpressurization of the mobile phase container can be very dangerous.

Best Regards,

Michael Hinsberg
http://www.hinsbarlabs.com


Posting is currently disabled in this topic. Contact your discussion moderator for more information.