Retention time variation

Chromatography Forum: LC Archives: Retention time variation
Top of pagePrevious messageNext messageBottom of pageLink to this message  By prabhakar on Thursday, December 28, 2000 - 08:19 pm:

Dear friends, Is there anybody who can enlighten me on what could be the RT variation criteria in HPLC system.


Top of pagePrevious messageNext messageBottom of pageLink to this message  By Tom M. on Friday, December 29, 2000 - 01:12 pm:

Hi,

I would start with the specification for yor particular make and model of HPLC. For example the current HP(Agilent)1100 series pumps have a spec of 0.3% RSD based on direct flow measurement while the older 1090 has a spec of 0.3% RSD based on retention time. Good Luck.


Top of pagePrevious messageNext messageBottom of pageLink to this message  By Cliff Woodward on Tuesday, January 2, 2001 - 08:37 am:

I am an Agilent employee. The criteria you need depend on your application. I have used peptide mapping for many years and I pick peaks based on retention time only as an initial identification. For this you need a variation of less than or equal to 0.3%. The resolution acheivable by 25cm columns is more of a limit than the variation in the Rt due to the pumping system. The criteria do not need be as stringent as this when you can use LC/MS, because MS is an absolute detector and can discriminate among compounds in the same chromatographic peak.