We have a HP 1050 with a 4.6 X 250 cm C18 column. All chromatograms have baseline noise a the begining we have tried a back pressure regulator and using the mobile as our solvent to no avail.
Any sugestions?
![]()
![]()
![]()
![]()
By Mike on Monday, June 21, 1999 - 03:15 pm:
Can you describe the "noise" ? also, is there any sign of pressure fluctuations ? I have seen this before on Beckman instruments with water to acetonitrile gradients and have found that microscopic bubbles in the flow cell are the cause. I would also try cleaning the check valves if you're seeing pressure changes.
![]()
![]()
![]()
![]()
By Thomas Pless on Tuesday, June 22, 1999 - 02:22 am:
If using TFA in the buffers and the disturbance is at the beginning of the gradient and appears as wavy baseline it depends on interactions between TFA and the gel. I have seen this on many chromatograms. This problem is mostly a visual disturbance and is not effecting the separation very much. Test to run without THF in buffer A. If the wavy baseline disappear I probably have guess right. You then have to try this:
1.lower TFA level in buffer A (to 0.060%, because you should not run without it)
2.test with 3% or 5% ACN in buffer A
3.Use wave length to 215 nm instead of 214 if possible (I do not no what you using today)
![]()
![]()
![]()
![]()
By John E. samela on Tuesday, June 22, 1999 - 10:48 am:
The chromatogram has a negitive peak then a positive spike. The pump module says the pressure is stable. We are runing ACN:Water 70:30 to 99:1 in 15 min.
![]()
![]()
![]()
![]()
By B.Buglio on Tuesday, June 22, 1999 - 07:15 pm:
What wave length are you monitoring at? Its not unusual to see baseline disturbance at the start of a gradient run if your monitoring at low (say 220nm) wavelength. Its a refractive index thing.Also try equilibrating the column longer before starting your next run. You could be seeing residual mixing of your end mp with beginning mp.
![]()
![]()
![]()
![]()
By SRS on Wednesday, June 23, 1999 - 02:34 am:
I hate to ask the dumb question but....What time is this happening at? could it be the peak from void volume? Also, maybe a posttime is needed based on what B.Buglio said (above) if there is a residual mixing problem.