Lofepramine from serum

Chromatography Forum: LC Archives: Lofepramine from serum
Top of pagePrevious messageNext messageBottom of pageLink to this message  By Phill on Thursday, January 18, 2001 - 04:13 am:

Has anyone tried extracting lofepramine from serum ? Given it's structural similarity to imipramine, it should have extracted easily using our existing liquid - liquid TCA method (as imipramine does), but it's very poor in comparison. I've also tried SPE using a mixed mode cartridge, again the other TCAs extract but not lofepramine. Any ideas would be welcome.


Top of pagePrevious messageNext messageBottom of pageLink to this message  By SOL on Wednesday, July 18, 2001 - 07:33 am:

The LLoQ for the following was 2.5 ng/ml. I hope it is of some help.

A 100 µl aliquot of sodium borohydride solution was added to 1 ml of plasma standard or sample in a disposable plastic tube and vortexed to mix. The tubes were incubated at room temperature for 15 minutes. The reaction was stopped by the addition of 1 ml of internal standard solution (93 ng/ml imipramine in 0.1 M HCl).

The extraction procedure was carried out automatically using a Gilson ASPEC. Varian Isolute columns (1 ml capacity, containing 100 mg cyanopropyl sorbent, end capped) were conditioned with 1 ml of methanol followed by 1 ml of 0.1 M HCl. The whole derivatized samples (2.1 ml) were applied to the columns and blown through under pressure. The columns were washed with 1 ml of water and the amino alcohol derivative of lofepramine and the internal standard were eluted with a mixture of methanol in HPLC mobile phase (30/70 v/v). 20 ìl of this solution was injected onto the HPLC column for chromatography.

Equipment and Operating Conditions

Mobile phase: Acetonitrile : methanol : 0.02M hydrogen phosphate buffer (55 : 5 : 40) - the pH was adjusted to 7.3 with sodium hydroxide
Flow rate: 1.0 ml/min
Temperature: 59C
Injector: Gilson ASPEC
Run time: 11 min (including extraction time)
Volume: 20 µl
Analytical Column: Spherisorb Cyano 5 µm (150 x 4.6 mm)
Detector: Coulochem 5100A with a 5011 Analytical cell
Output: 10 mvfsd
Ranges: Electrode 1 0.50 V
Electrode 2 0.80 V
Gain: Electrode 2 50 x 100 µA
Filter: 10 seconds
Auto zero: -15%


Posting is currently disabled in this topic. Contact your discussion moderator for more information.