Hi ALL,
Can you please comment on the OQ/PQ validation tests for the pump module in a HPLC system. The tests followed currently are Flow accuracy, pressure hold test and.. I need to know what all other tests to be incorporated in the OQ documents.
thanks
![]()
![]()
![]()
![]()
By Anonymous on Monday, March 19, 2001 - 11:04 pm:
if the pump is used to mix, the accuracy off the mixing....
![]()
![]()
![]()
![]()
By Kaf on Tuesday, March 20, 2001 - 12:51 am:
noise: you should regist the evolution signal/noise evolution in a isocratic system
![]()
![]()
![]()
![]()
By Tom on Tuesday, March 20, 2001 - 04:26 pm:
It depends on how you use the pump. If you have use online mixing you should test the proportioning accuracy. If you run gradients you should perform a gradient proportioning test. I recommend using a digital flow meter and you can test for both flow accuracy and precision. If you are going to be performing a gradient proporting test anyway you might as well configure it to give you dwell volume data as well. Good Luck.
![]()
![]()
![]()
![]()
By mscdak on Tuesday, March 27, 2001 - 06:53 pm:
I conduct instrument validation for laboratories that need to be in compliance with regulations. For Pumps I usually run a Leak Test(this is a Diagnostic Test),Pressure Test,Flow Accuracy Test, Noise Test,Gradient Linearity, Gradient Step Accuracy and etc...For more information call me at BIOMATIC TECHNOLOGIES INC. 781-341-1444.
I will be happy to answer any of your validation questions
Michael Collins- Manager of Validation and Chemistry
![]()
![]()
![]()
![]()
By Anonymous on Thursday, May 3, 2001 - 08:11 am:
How the flow accuracy test has to be done? Is it with the column on the line or without it?
![]()
![]()
![]()
![]()
By Merlin on Friday, May 4, 2001 - 02:34 pm:
Flow accuracy is best measured by timing how long it takes to fill a 10 mL volumetric flask to the mark with water at 1 mL/min. If you want 2% accuracy, it should take 9.8-10.2 minutes. As pumps work best under pressure, you need some form of back pressure device. You can leave the column in (assuming it's an RP column), but I like to use a 2m length of 0.005" id tubing. This gives about 450 psi at 1 mL/min of water. It might also be a good idea to try different solvents, to see how solvent compressibility affects the accuracy.
![]()
![]()
![]()
![]()
By Anonymous on Tuesday, May 15, 2001 - 04:51 pm:
Considering the accuracy of some of today's pumps, you can use then to calibrate your graduate!!!