I am using isocratic HPLC (UV detection) to quantify chloride, dextrose, and acetic acid using 0.01 N H2SO4 as mobile phase. Method is well established and has worked well for many years. All of a sudden, after a column reconditioning, the chromatography is awful. The first peak, chloride, developed a huge tail and now the following peak, dextrose, is not resolvable. I've changed mix filter, guard column, and column. Does anyone have any further advice for troubleshooting this method?
![]()
![]()
![]()
![]()
By tom jupille on Thursday, April 12, 2001 - 10:38 am:
Need more details:
- from the mobile phase, I would surmise you're using a cation exchange column in the H+ form for the separation. What, exactly, did you do to "recondition" the column?
- I interpret your post to say that the problem continues even after you replace the analytical column with a new one. Is this correct?
- What wavelength are you using?
After all that, I'll go out on a limb and speculate: if the problem cropped up abruptly after your disconnected/connected your column to recondition it, and if the problem persists with a new column, the most likely cause is a mis-connected fitting or excessively large piece of tubing between injector and column or between column and detector.
-- Tom Jupille / LC Resources Inc.
Posting is currently disabled in this topic. Contact your discussion moderator for more information.