This's been on going for months: I have to change the frit in the purge valve every week! With the purge valve opened, 5ml/min pressure will go above 30 bar after roughly a week of medium use (~150 injections)
I am using regular phosphate and acetate buffers at 25 - 50mM range. Attempts to fix the problem: Change plungers, pump seals, purge lines with 50/50 water/IPA and clean resorviors. What else could be the cause of the problem? Please help!
![]()
![]()
![]()
![]()
By Anonymous on Wednesday, June 20, 2001 - 01:30 pm:
Phosphate buffer on chn A and solvent on channel B. The purge valve is open when you are are priming these lines at conc. that ppt the buffer.
![]()
![]()
![]()
![]()
By David Blais on Friday, June 22, 2001 - 07:47 am:
You could look into one of these two things, also. First, to get to the purge valve frit, you have to remove a little plastic cap with a thin gold disc; if that disc is scored, damaged, etc., you could see a dramatic increase in pressure while purging - replace it. Second, you simply might have a bad purge valve altogether. Order another valve from Agilent and see if that makes a difference. If not, call Agilent Tech Support for and pick their brains. Hope this helps.
![]()
![]()
![]()
![]()
By craig on Friday, June 22, 2001 - 08:35 am:
I had similiar problems before, and I agree with
what David Blais said above, but try and sonicate
the purge valve in acetonitrile for about 10 minutes. That and a methylene chloride flush of the system cleared the problem up for me.