Non-zero y-intercept for Linearity of HPLC Assay

Chromatography Forum: LC Archives: Non-zero y-intercept for Linearity of HPLC Assay
Top of pagePrevious messageNext messageBottom of pageLink to this message  By K. Kolodsick on Tuesday, July 17, 2001 - 07:55 am:

I have been working on an HPLC assay (C18 100 x 4.6 mm, 3 micron - mobile phase 40%ACN : 60% 25mM pH7 phosphate) with UV detection (270 nm).

When I run linearity standards from 50-150% of the assay concentration I get a significant y-intercept from my linearity plot (2.5% of the 100% response).

What are some common causes of this type of linear curve? My standards are prepared in 40:40:20 MeOH:ACN:H20 and injection volume is 10 mcL. Peak shape is good and blanks show no peaks which interfere with the analyte of interest.

Also of interest is that if I prepare standards of the same concentrations mentioned above by spiking into placebo samples, I get a very small intercept (0.3% of the 100% response).

Any suggestions?

Thanks in advance


Top of pagePrevious messageNext messageBottom of pageLink to this message  By Daren on Tuesday, July 17, 2001 - 09:32 am:

I would run your calibration curve down as low as you can go, maybe around 5 or 10 % of your target conc. with a few points between there and 50%. Ideally you will be able to show that the lower you bring the calibration curve the more the y-intercept approaches zero. If you can show that this is the case then you can force your line through zero. Not sure about the placebo, I would assume it is being separated away from your std on column. The main in thing is that the slope is consistent more than the y-intercept.


Top of pagePrevious messageNext messageBottom of pageLink to this message  By Uwe Neue on Tuesday, July 17, 2001 - 05:39 pm:

I think that you may be outside the upper linearity range of your detector. At the upper end, the signal goes down. If your upper calibration signal is anywhere close to 1 AU, this is the problem. If you are not there, it is still possible that the whole thing is caused by a dirty cell window, or a low lamp intensity.


Top of pagePrevious messageNext messageBottom of pageLink to this message  By DR on Wednesday, July 18, 2001 - 06:39 am:

If Uwe's first suggestion is correct, you should be able to see it by comparing the shape of the tops of peaks - the high concentration samples (if non-linear) will appear to be different from peaks that are in the linear range of the LC - specifically, check for plateaus. Excessive fronting as concentrations rise is indicative of too little column capacity for the amounts of analyte being injected...


Top of pagePrevious messageNext messageBottom of pageLink to this message  By K. Kolodsick on Wednesday, July 18, 2001 - 10:35 am:

The signal tops out around 200 mAU, so I don't think that I am exceeding the linear range of the detector. The peak shapes are consistent through the entire range I studied.

I checked the lamp hours on Uwe's suggestion, and indeed, this lamp is pretty old. I have set up a run on a second instrument where the lamp was recently changed. Perhaps this was the problem all along.

Thanks for your input.


Top of pagePrevious messageNext messageBottom of pageLink to this message  By Anonymous on Saturday, July 21, 2001 - 11:53 am:

Non-linearity can occur long before the detector is saturated and produces a plateau.

Any of the "normal" deviations from Beer's Law may be present as analyte concentration increases. Then, regardless of the analyte, the detector will have a linear range due to its design.


Top of pagePrevious messageNext messageBottom of pageLink to this message  By K. Kolodsick on Monday, July 23, 2001 - 05:12 am:

Indeed the run repeated on a different instrument with a recently installed lamp gave a "zero" intercept.

Thanks for your suggestions.