I have a method for the analysis of impurities related to 1 of 3 active ingredients in a drug product. One of the other active ingredients adheres to the column. Unless I clean the column on 100 % Methanol after any sample injections for 1-2 hours, I will get carryover. I have been able to demonstrate that the carryover is not due to the LC. On occaison we are seeing a significant impurity peak which we have determined is a degradant of the active ingredient that is adhering to the column.
I am using a 5-µm 4.6x150-mm Zorbax Eclipse XDB-C18. The mobile phase is methanol: pH 3.0 25mM sodium phosphate, monobasic,(20:80). Would a guard column help this situation?
![]()
![]()
![]()
![]()
By jclark on Thursday, November 1, 2001 - 01:26 pm:
A guard column would only help temporarily. You need to do some sample cleanup, probably by SPE.
![]()
![]()
![]()
![]()
By Uwe Neue on Thursday, November 1, 2001 - 04:14 pm:
Think about why this active ingredient sticks! Most likely, it is more hydrophobic, and a simple wash with a higher % organic will solve your problem. I personally would do it after every single run, but you may get away with doing it less frequently.
I can not see why you would need to wash the column for 1-2 hours in methanol. It may be that the problem compound is retained for other reasons than hydrophobic interaction, and washing with 50/50 methanol-buffer would do a better jop than washing with methanol. If indeed the interaction is hydrophobic, you could wash the column with acetonitrile or THF after each run.
Posting is currently disabled in this topic. Contact your discussion moderator for more information.