Hello everybody!
I´m having a problem with my high speed columns :everyone I´ve tried, produced band broadening and tailing peaks after a few runs.
Now I´m with a microbore one of 5cm x 2.1mm, of 3.5um, C18.I got a flow of 0.4mL/min, maximum; and never had preasure problems .(Never more than 150bar)
The question is if 0.4mL/min is too much flow.Could it have comprimed the silica packing and generated a void volume?
Thank you very much.
![]()
![]()
![]()
![]()
By tom jupille on Friday, May 3, 2002 - 03:51 pm:
0.4 mL/min through a 2.1 mm id column is equivalent to 1.9 ml/min through a "standard" 4.6 mm id column (proportional to the square of the column diameter). Given a length of 5 cm, this flow rate should not damage a 3.5-micron packing.
If you see band broadening and tailing all the time, then I would suspect extra-column volume: injection volume, connecting tubing length/diameter, fittings, and detector cell volume.
I'm assuming that you are taking the usual precautions:
- injection volume down around 1 or 2 microliters
- 0.005" id connecting tubing
- low volume (1-2 microliters) flow cell.
If your efficiency looks OK for a "few" runs (what is "a few" in this case? 3? 10? . . .) and then drops off, I would start off by double-checking my fittings to make sure something hasn't come loose. Then I'd look at my sample chemistry to see if I'm not precipitating something.
![]()
![]()
![]()
![]()
By Curbiro on Saturday, May 4, 2002 - 08:17 am:
Hello Tom,
Thanks for answering so fast...
My English is a little bit limitated, so perhaps I´m going to look too much direct.
I´m filtering water, AcN, and KH2PO4 dsn, and sample, trough 2um; also I´m using an adittional microfilter pre-column of 2 um.
The equipment is being shared with other colleague, who is also working in hs-hplc and she doesn´t have any problems, so I think that fittins, flow cell or tubing connections are not the problem...
I´m ingecting 0.5uL.
I wash every day the column at the end of working, to be sure that nothing is retained .
Really, it has happend to me with another conventional column too(I didn´t supered 2mL/min).
So I´m thinking about the possibility of a bad packing system...
![]()
![]()
![]()
![]()
By Anonymous on Saturday, May 4, 2002 - 08:58 am:
bad packing system? Do you pack the columns yourself?
![]()
![]()
![]()
![]()
By Curbiro on Saturday, May 4, 2002 - 01:12 pm:
No, of course, not.
But it´s been a year with a lot of problems with the columns from the same company.
I cant find other explanation....Do you?
![]()
![]()
![]()
![]()
By A.Nonymous on Sunday, May 5, 2002 - 10:59 am:
How do you wash your columns, what kind of samples do you inject?
Is you wash strong enough?
![]()
![]()
![]()
![]()
By Curbiro on Monday, May 6, 2002 - 02:42 am:
I wash the column everyday .I start with water and 5%AcN for almost half an hour; after, I go to 100% AcN and I´m there for 10 to 15 minutes and at the end, I equilibrate the column in 65 %.
It´s a C18.
But today, I discobered that if I put 100%AcN after the washing, something eluted...
Do you know other way of washing better?
Thanks...
![]()
![]()
![]()
![]()
By Curbiro... on Monday, May 6, 2002 - 02:43 am:
I inject basic antibacterials...
![]()
![]()
![]()
![]()
By A.Nonymous on Monday, May 6, 2002 - 11:36 am:
Have you tried a wash with THF already, this sometimes may help
![]()
![]()
![]()
![]()
By Anonymous on Monday, May 6, 2002 - 04:26 pm:
In one of the 'HPLC Troubleshooting' columns in the magazine there was an instance of a chemist using PEEK tubing that 'slipped' under high pressure, generating a void volume after a few injections.
I have hundreds of runs on a column that runs over 200 bar, but I have metal tubing on the inlet side--no possibility of slippage.
There are many possible reasons, but this one is easy to eliminate.
![]()
![]()
![]()
![]()
By Anonymous on Monday, May 6, 2002 - 04:27 pm:
If junk from your injections is a suspected problem, how about using a guard column?
![]()
![]()
![]()
![]()
By Curbiro on Tuesday, May 7, 2002 - 06:53 am:
Every conection is of steal; The only thing I use of peek are fingertight of high preasure...
I´m trying with a precolumn..
Thank to everybody..
Regards.
Posting is currently disabled in this topic. Contact your discussion moderator for more information.