2-methyl-4-isothiazolin-3-one (MIT)

Chromatography Forum: LC Archives: 2-methyl-4-isothiazolin-3-one (MIT)
Top of pagePrevious messageNext messageBottom of pageLink to this message  By Anonymous on Thursday, November 7, 2002 - 11:32 am:

I'm trying to determine the amount of preservative (MIT and MIT-Cl) left in product at the end of shelf life. I am unable to get good peak seperation with the method I've found, mobile: 50% mehtanol 50% water at 1.3 mL/min, c18 column, 275 nm. My standard seperates nicely but I can't seem to get the peaks of my sample far enough away from everything else that elutes before the MIT. I'm very new to HPLC and wolud welcome any suggestions. I appreciate any help you can offer.


Top of pagePrevious messageNext messageBottom of pageLink to this message  By Anonymous on Thursday, November 7, 2002 - 11:47 am:

Try decreasing the amount of organic in your mobile phase. Try using 60% water and 40% methanol. A rule of thumb in RP chromatorgraphy decreasing the amount of organic will increase the resolution between the two peaks of interest.

Is this a validated method and are you working in a GMP environment? If so, you can't change the method you are using without re-validating it.

Have you tried filtering your sample to get rid of the "junk"?

Hope this helps.


Top of pagePrevious messageNext messageBottom of pageLink to this message  By Anonymous on Thursday, November 7, 2002 - 01:38 pm:

We use the following method for analysis and got just enough retention for MIT to be away from other non-retain compounds. The key is to use longer column. We used 150mm column and was not able to get enough retention for MIT. Try this method.

HPLC column: Waters uBondpack C18, 5 um 3.9 mm x 300 mm
Detector: Waters 996 PDA, 275 nm
Mobile phase: 50/50 methanol/water


Top of pagePrevious messageNext messageBottom of pageLink to this message  By Anonymous on Friday, November 8, 2002 - 12:33 pm:

We do this assay all the time, in many different consumer product matrices including surfactants.
10 cm x 2.1 mm i.d 5u RP-18 column coupled with 3 cm x 2.1 mm i.d. 5u RP-18 guard cartridge
Solvent A = 0.4% acetic acid in H2O, Solvent B = HPLC CH3OH, Flow = 0.25 ml/min, Detector Wavelength = 273 nm, Column temperature = 27 *C, Run time = 17 minutes, Post-time = 5.0 minutes, Injection volume = 10.0 ul

HPLC Pump Gradient Conditions

Time (minutes) Event
Initial value 13% CH3OH
5.0 min. 40% CH3OH
7.0 min. 40% CH3OH
8.0 min. 100% CH3OH
13.8 min. 100% CH3OH
15.0 min. 13% CH3OH
17.0 min. stop run
Post-time 5 minutes for re-equilibration.
With these conditions the retention time for methylisothiazolinone was approximately 3.3 minutes and the retention time for methylchloroisothiazolinone was about 9 minutes.