On a Discovery HS F5 (150 x 4 mm) I noticed some strange behaviour when analysing three structurally similar neutral compounds under isocratic conditions:
Resolution (Rs=(2(RT2-RT1))/(W2+W1) ) changes with the flow rate. At 1 ml per min the second and third peak are just shoulders on the first peak. At 0.75 mL/min all three peaks are separated (Rs 2/1 = 2.4, Rs 3/2 = 1.3) and at 0.5 mL there is nice baseline separation (Rs 2/1 = 3.6, Rs 3/2 = 1.9).
Theoretically the flow rate can influence only the number of theoretical plates via the van-Deemter equation. With our example for the first peak efficiency increases with decreasing flow rate: N=6000 for 1 mL/min, 9700 for 0.75 mL/min and 12000 for 0.5 mL/min.
However, that doesn’t explain solely the improved separation at low flow rates.
Also k' is changing (for peak 1 1.9 at 1 mL/min, 3.4 at 0.75 mL/min and 5.3 at 0.5 mL/min). The alpha-value increases slightly for the second peak (1.11, 1.13 and 1.16) and is stable for the third.
I do not understand the increasing k’ with decreasing flow rate. Has someone experienced something similar? Has some one a explanation for this? Are there any articles about retention behavior on PFP columns?
Alex
![]()
![]()
![]()
![]()
By Anonymous on Wednesday, August 13, 2003 - 05:05 pm:
The change in retention factor is the only thing that is strange. What is the compound with which you measure t0? Is V0 (t0* flow rate) constant?
![]()
![]()
![]()
![]()
By Alex on Thursday, August 14, 2003 - 03:54 am:
Maybe not strange, but still unexpected is the increase in efficiency at low flow rates. With a 4 mm column and 3 µm material I would expect the highest efficiency around 1 mL / min, slowly decreasing with higher flow rates and more pronounced decreasing at flow rates <0.4 mL /min.
t0 is estimated, assuming a constant V0. If i take a minor disturbance in the chromatogramm as marker V0 is decresing with lower flow rates (at 0.5 mL/min the disturbance is observed at 3.12 min, for calculations 3.9 min (2x t0 at 1 mL/min).
When using the disturbance as t0 marker I would get even higher k' values for the lower flow rates.
I should try uracil or acetone as t0 markers
Thanks for the comment,
Alex
![]()
![]()
![]()
![]()
By Anonymous on Thursday, August 14, 2003 - 08:32 pm:
There is nothing extraordinary with a platecount of 12000 for a 15 cm 3 micron column. It is actually not up to standard. A decent manufacturer should supply higher numbers on a routine basis. Why do you expect the high platecount at a higher flow rate? What is your mobile phase?
![]()
![]()
![]()
![]()
By Alex on Monday, August 18, 2003 - 06:34 am:
N= 12000 is sufficient and realistic, especially if you do not work with 2 µL injections and selected compounds. Solvent is methanol / water (50/50).
Again: are there any publications on retention behaviour on PFP columns?. Has anyone observed change of k' values with flow rate?
![]()
![]()
![]()
![]()
By Anonymous on Monday, August 18, 2003 - 04:06 pm:
It is not very likely that you are really looking at change in retention factor. I suspect that you are looking at a change in t0. Of course you won't know until you use a real t0 compound to measure it. There could be leak or airbubble during injection, or there could be a delay when your machine starts the chromatogram. I have seen both.