Problem: linearity variation of instrument

Chromatography Forum: LC Archives: Problem: linearity variation of instrument
Top of pagePrevious messageNext messageBottom of pageLink to this message  By May on Tuesday, December 23, 2003 - 07:58 am:

Hi, there,

I am doing a HPLC method validaton (gradient, UV detection). I am using 2 same types of instruments for the validation. One instrument consistently gives perfect linearity for response vs standard concentrations, but another instrument behaves inconsistently when checked with the same standard solutions: sometimes the linearity is ok, sometimes very bad. I am wondering what's the possible cause of the problem.


Top of pagePrevious messageNext messageBottom of pageLink to this message  By laurent on Tuesday, December 23, 2003 - 11:16 am:

The causes could be wide.
Change elution solvents (degas), clean solvent filters, check the pump flow (and maybe change seals), check also the injection part (seals too and injection volume precision).
If not resolved then time to find out the old troubleshouting guide, remove the dust and check all the possibilities :-)


Top of pagePrevious messageNext messageBottom of pageLink to this message  By hk on Thursday, December 25, 2003 - 06:17 pm:

Agree with Laurent that the possibilities could be numerous. Check your user manual for possible causes, or call the manufacturer. Anyway, at first glance, your problematic LC's faulty part(s) could be isolated in the injector area (if any of these apply to your instrument) i.e. injector seal (leaky), faulty injector metering motor, plugged injector, switching valve, or the old plunger. I hope this help.


Top of pagePrevious messageNext messageBottom of pageLink to this message  By Anonymous on Friday, December 26, 2003 - 03:17 am:

maybe the lamp of your second detector is at the end of its lifetime


Add a Message


This is a private posting area. A valid username and password combination is required to post messages to this discussion.
Username:  
Password: