The problem began when the system ran dry. I added fresh mobile phase solution to the system, primed the pump and finally got some flow going through my RI detector out waste line. I also purge the flow cells, just in case there was a possibilty of air in the cells. Afterwards I noted that my baseline was drifting (descending).
I have tried to determine if the problem is with my Waters 510 pump (note: I had previously changed the seals and cleaned both inlet and outlet chk valves for both pump heads) or the Waters 410 detector. I monitored the baseline w/o flow. Instead of a straight line, I get a noisy baseline. Is there a possibiltiy that there maybe an electrical problem or something else I haven't considered.
![]()
![]()
![]()
![]()
By Anonymous on Wednesday, January 7, 2004 - 04:21 pm:
If there is air in the cell, you probably won't get it out by pumping mobile phase through it. It would be a good idea to pump 100% MeOH through the cell and then purge the reference side with the MeOH. It will take lots of time to get it to settle back down, so have some patience.
![]()
![]()
![]()
![]()
By J.P. on Wednesday, January 7, 2004 - 08:57 pm:
Is Your mobile phase degased? If not, wavy baseline is typical.
![]()
![]()
![]()
![]()
By Durk on Thursday, January 8, 2004 - 07:22 am:
We flush the reference channel with 2-propanol during night (low flow) and after that the sample channel at low flow until the baseline is stable. The possibility is that a stable baseline is achieved after 1 or 2 days.
![]()
![]()
![]()
![]()
By Anonymous on Sunday, January 11, 2004 - 11:55 pm:
Try to generate some back pressure on outlet tube of detector , when flow( IPA)is on. Use your finger to block the flow and than let is go. This will remove the air from the flow cell.