Vitamin D3

Chromatography Forum: LC Archives: Vitamin D3
Top of pagePrevious messageNext messageBottom of pageLink to this message  By Anonymous on Tuesday, January 27, 2004 - 11:25 am:

I am developing an assay for purity of Vitamin D3 - cholecalciferol. Currently I am using a 150mm x 4.6mm, 5um C18 column with Methanol/water as mobile phase. I've tried gradients and isocratic in ratios from 80:20 MeOH:water to 100% water. The problem is is that there appears to be some sort of impurity, a minor peak co-eluting with the cholecalciferol peak. I am unable to resolve the two, changing the mobile phase only seems to effect retention time.

I am using a UV PDA detector, the minor peak has a very similar UV spectrum to the main peak.

I am not familiar with this analyte, I've never done vitamin D before. Does anyone have any suggestions as to what this peak may be, how to resolve it etc? Bear in mind that I am more of a GCMS guy than HPLC, so please speak slowly and avoid big words.

Also, how toxic is this material, I was a little intimidated by the MSDS...


Thanks,

Mike


Top of pagePrevious messageNext messageBottom of pageLink to this message  By Einar Pontén - SeQuant AB on Tuesday, January 27, 2004 - 04:08 pm:

Mike,

It's "easy". You only need to keep three things in mind. Efficiency, capacity, and selectivity

If you have an efficient system expressed as a large plate number then bandbroadening within and outside the column is okey as well as flow rate.

While changing the mobile phase composition stepwise or in a gradient run then you mainly control the capacity factor (retention).

Finally, the whole great world of chemistry can be utilised to influence the selectivity. I think you should focus on that part.

For instance, try to use a 5-20 mM buffer salt in the eluent at different pH values (2-7 on silica column), try acetonitrile as organic modifier and so on..

Good luck!


Top of pagePrevious messageNext messageBottom of pageLink to this message  By Anonymous on Friday, January 30, 2004 - 06:05 am:

USE
75% ACETONIRILE AND 25% METHANOL ON C18 250 CM 5 um column you will get nice one at 17 minutes which is stability indicating or even you can use BP method for analysis


Top of pagePrevious messageNext messageBottom of pageLink to this message  By Anonymous on Thursday, February 5, 2004 - 12:59 pm:

Is it a synthetic analog of D3 or a real D3?
Also keep in mind that D3 is not water soluble, so 100% water is not a suitable mobile phase for it or its related compounds.


Top of pagePrevious messageNext messageBottom of pageLink to this message  By Anonymous on Wednesday, March 17, 2004 - 01:23 pm:

Mike,

The interfering peak is probably vitamin D2 (ergocalciferol) or pro-D3. Both of these will have spectra pretty much identical to vit D3. Most reverse-phase columns do not give very good resolution of D2 and D3. One column that does is the Vydac 210TP54 column. We use a 25 cm column with mobile phase of 90:5:5 Acetonitrile:Methanol:Water at 1.5 mL/min. This should give you good resolution of the peaks and allow you to quantitate with appropriate standards.

-Ken