hi everyone!
I am using the turbochrome software of a perkin elmer HPLC, when I select the wavelenghts I want to use, I should also choose the reference wavelenghts...what is their function and what are the criteria to choose them?
thanks
annav
![]()
![]()
![]()
![]()
By Henrik Vogelius on Thursday, May 20, 2004 - 01:25 am:
Choice of a proper reference is effective in reducing the baseline drift during gradient elutions. Without a reference, the chromatographic signal drifts upward
or downward due to changes in the mobile phase induced by the gradient. The drift is almost completely eliminated by use of a reference signal. wavelenght should be close to the sample wavelenght and try a bandwith for about 50nm for this wavelenght. but the ref-wavelenght may not overlab the samplewavelenght.
![]()
![]()
![]()
![]()
By Anonymous on Thursday, May 20, 2004 - 08:54 am:
On a DAD, all of the light from the source lamp(s) is directed through the flowcell. By using a reference, changes in the lamp intensity as well as changes in MP absorbance can be negated from the sample signal.
In a variable wavelength detector, the lamp energy is split, so that there a 2 discreet photodiodes for the signal to be interpreted.
![]()
![]()
![]()
![]()
By Anonymous on Thursday, May 20, 2004 - 10:13 am:
Sorry, do you mind telling me why the reference wavelength should be close to sample wavelength.
Thanks so much!!!
jin
![]()
![]()
![]()
![]()
By Henrik Vogelius on Thursday, May 20, 2004 - 10:44 am:
You use the reference, to subtract the background at the sample wavelength doing the run, and therefore it should act as much as the sample wavelength.
![]()
![]()
![]()
![]()
By Annav on Monday, May 24, 2004 - 02:32 am:
how close to the sample wavelength should be the ref-wavelength? should it be higher or lower?
my sample has its UVmax at 230 nm, I am using acetonitrile as mobile phase, I choose 210 nm as ref-wavelenght but I can not see any peak.with another sample I had, chlorogenic acid, I used 325 nm and 360 nm as reference wavelength I did not see any peak, when I used 220 as ref-wavelength I had the peak(mobile phase MeOH)...
I am a little bit confused...
![]()
![]()
![]()
![]()
By Henrik Vogelius on Monday, May 24, 2004 - 05:49 am:
Hi
What's also important is the bandwidth for this ref wavelength. If it overlaps the sample wavelength you don't see any peak or maybe just a small peak. If you look at the iso-plot for your sample place you ref wavelength outside the peak, wher you have a low absorbans. I will suggest ref wavelength = 500 at 50 nm bandwidth. You can try to move it closer to the 230 nm, but don’t overlap, so minimum 300 nm.
![]()
![]()
![]()
![]()
By annav on Monday, May 24, 2004 - 08:41 am:
thanks very much for the suggestion!
I will try in this way!
![]()
![]()
![]()
![]()
By A.Mouse on Tuesday, May 25, 2004 - 06:21 pm:
I don't get it. If you are using a reference wavelength close to the sample wavelength, you are suppressing signal. You said this yourself - if you select the reference wavelength too close to the sample wavelength, the signal goes away. Don't you need to select a reference wavelength way away from the measurement wavelength, preferentially at a wavelength where the signal of the sample is rather small? How small does the signal need to get to maximise signal/noise ratio?
![]()
![]()
![]()
![]()
By Henrik Vogelius on Wednesday, May 26, 2004 - 02:41 pm:
Yes I will suggest that you start far away from the quantification wavelength, look at the S/N ratio, the peak height and the drift, then move it closer to the quantification wavelength and take another look. By doing that, you are optimising ref- wavelength. So you see it a question on how dose your solvent absorb at a high wavelength, I think not much compared to the quantification wavelength, therefore you use a high bandwidth to let in more energy for the photo multiplier. And remember this ref wavelength is the ref point for the measured signal, so every time you have a data point you also have a autozero, and therefore there is no drift from the gradient. On a simple UV-detector you do the autozero at the beginning of the run, with a ref energy point measured from a split light before entering the flowcell, and therefore it can not measure baseline movement caused by the solvent.
Hope you not more confused now!