I am using 0.1%TFA in HPLC water and 0.1%TFA in HPLC MeCN with a gradient 10%B to 40%B in 4mins and a C18 BDS column/215nm.
I have an inevitable ramp up in the baseline while the composition changes (I assume slight different concs of TFA are causing this).
I measured the UV abs at 215nm today of each solvent and worked out a dilution of the 0.1%TFA in MeCN with MeCN to get the same abs as the 0.1%TFA in water (9parts to 1part).
I will try this experiment tomorrow - has anyone tried this before? Did it work for you?
![]()
![]()
![]()
![]()
By BT on Thursday, August 26, 2004 - 10:17 am:
Your ramp is due to the shift from water to acetonitrile and the differences in their UV absorbances. Most solvents have greater absorbance than water, so pretty much every gradient will do this, especially at lower wavelengths. To confirm, run your gradient with just MeCN and water. You'll probably see the same baseline.
![]()
![]()
![]()
![]()
By A on Thursday, August 26, 2004 - 10:27 am:
A lot of people doing this type of gradient try to balance the TFA to get less of a rise in the baseline. Often it is 0.1% TFA in water against 0.085% TFA in the ACN. This can help.
![]()
![]()
![]()
![]()
By WK on Friday, August 27, 2004 - 02:50 am:
Thank you BT and M.
Update:
My 9:1 dilution still had a baseline rise - not so bad (maybe 75%) compared to the undiluted rise.
And the change in TFA conc in MeCN had a small effect on retention.
I dont want to go back to phosphate - I will be forever unclogging/replacing capillaries!!
My next experiment will be to try and get the baseline flatter and then change the ramp to resolve my close eluting pairs.
Watch this space!
![]()
![]()
![]()
![]()
By BT on Friday, August 27, 2004 - 07:40 am:
Are you using the increase in absorbance from the MeCN and the decreasing absorbance of the TFA (due to lowered concentration in the B mobile phase) to end up with a totally flat baseline? I might have to try that with a different buffer that is giving me problems.
![]()
![]()
![]()
![]()
By WK on Friday, August 27, 2004 - 08:05 am:
Yes,BT.
I know that it probably won't work - but its worth a try! Something else in the separation will give - perhaps the buffering.
One of my peaks comes out on the "hill" and its not good integration.
I lose sensitivity at wavelengths above 215nm.
Let me know how you get on...
![]()
![]()
![]()
![]()
By Chris Pohl on Friday, August 27, 2004 - 08:13 am:
WK
The reason why your first attempt did not completely eliminate the problem is that there are two components to the drift: the difference in the absorbence of your starting and ending eluent and an added portion coming from the stationary phase as the increasing solvent decreases the amount of TFA adsorbed on stationary phase. As you've already seen, the first part is straightforward to correct for but in my experience it's quite difficult to completely eliminate the second effect because of the details of the gradient start and end. Usually, if you take the approach of making a compensating decrease in the absorbance of the strong eluent you can get the beginning and ending level to match but then you usually have a curved baseline with a high point in the middle of your gradient.
![]()
![]()
![]()
![]()
By WK on Friday, August 27, 2004 - 08:24 am:
Thanks Chris,
Changing the subject a little:
The TFA adsorbed onto the stationary phase - does it get removed upon neat MeCN washing or does it build up over time to saturate the column?
This is an interesting thread.
Next week I will try to see if I can improve so that my peak comes out on a flatish piece of baseline.
![]()
![]()
![]()
![]()
By Chris Pohl on Friday, August 27, 2004 - 08:38 am:
WK
Washing with neat MeCN should remove the adsorbed TFA quantitatively from a reversed phase column with the exception of any TFA which is interacting "nonspecifically" with polar sites in the underlying support material (residual silanols, metal contamination, trace residuals from samples, etc.).