I am using a cellulose based chiral column to separate the enantiomers of organophosphates on an HP 1100 system that is equipped with a diode array detector. I observe an increase in peak areas when the flow rate is reduced. One example, when the flow rate (100% hexane) was decreased from 0.8 ml/min to 0.3 ml/min the areas of the separated enantiomers increased by a factor of three (monitored at 205 nm). My sources tell me that UV response is relatively insensitive to flow rate, but to date I have observed the same phenomenon for four organophosphates. The injected amounts are in the lower range of the linear calibration lines that were determined for each organophosphate.
![]()
![]()
![]()
![]()
By Bruce Freeman on Tuesday, February 1, 2000 - 10:35 am:
UV response expressed as peak height is insensitive to flow rate. This is not necessarily so for UV response expressed as peak area.
Think of the peak as a volume of liquid emerging from the column and entering the detector. The detector cell contains a finite volume (much) less than that of the peak. If the peak flows slower through the detector, it is measured over a longer time. Hence, the peak area (which is usually expressed in units like volt-seconds) is greater when the flow rate is lower.
Assuming no change in column efficiency, the relationship should be direct. Hence in your case one would expect peak area to increase by a factor of 0.8/0.3 = 2.7, which is approximately what you report.
If you were to change the units of peak area from volt-seconds to volt-milliliters (by multiplying by the flow rate) you would observe little or no change in peak area due to change in the flow.