I am running a reverse phase method for estradiol using 55:45, ch3cn:water as the mp. My column is A YMC C18, 3.0 x 250 mm. particle size 5 ucron. column temp = 30c. I noticed the retention time and area count decreasing as the run continues. I am bracketing every six samples with standards. Does anyone know why the RT and area count are decreasing and should I be concerned about this?
Thanks-R
![]()
![]()
![]()
![]()
By Uwe Neue on Monday, March 6, 2000 - 07:29 pm:
The two observations together indicate a problem with the flow, i.e. the flow seems to be increasing over time. Check if this is indeed the case, then we worry about the cause.
![]()
![]()
![]()
![]()
By Anonymous on Tuesday, March 7, 2000 - 02:33 pm:
Check to make sure the frit (if any) on your water reservoir is clean. Too much ACN, assuming you are mixing on-line, will cause RT's to drop slowly.
![]()
![]()
![]()
![]()
By sha on Wednesday, March 8, 2000 - 10:34 am:
Do your sample contain other compounds? eg. PEG. Do you have the same problem if you only inject your standard six times? I would think your sample ingredient play a role.
![]()
![]()
![]()
![]()
By daren on Friday, March 10, 2000 - 06:13 am:
How old is your column? If you've checked your system and filters and they're fine, then I would try a new column. First thing I would do though is remove the mobile phase inlet filters and try the run with fresh ACN/H20. If that's the problem you can sonicate the filters in ethanol for a while.
![]()
![]()
![]()
![]()
By Jonathan on Tuesday, March 14, 2000 - 07:26 am:
Is your baseline on a downward slope? This can tend to shift retention time backward...you might want to check whether this is a occurring only at certain points in a run, i.e., what the conditions are surrounding this shift.